QUESTION IMAGE
Question
a radiographic technique chart recommends 4 milliampere - seconds (mas) and 68 kilovoltage peak (kvp) at 72 inches for a certain procedure. how will the receptor exposure be affected if the radiograph is performed using the recommended exposure technique but at 60 inches? receptor exposure will not be affected receptor exposure will be lower than desired receptor exposure will be higher than necessary
Step1: Recall the inverse - square law
The intensity of the x - ray beam (and thus receptor exposure) is inversely proportional to the square of the distance from the source. The formula is $I_1d_1^{2}=I_2d_2^{2}$, where $I$ is the intensity of the x - ray beam and $d$ is the distance.
Step2: Compare the distances
The recommended distance $d_1 = 72$ inches and the new distance $d_2=60$ inches. Since $d_2
Snap & solve any problem in the app
Get step-by-step solutions on Sovi AI
Photo-based solutions with guided steps
Explore more problems and detailed explanations
Receptor exposure will be higher than necessary