a portable chest x - ray is to be acquired on an emergency room (er) patient. the technique chart indicates the exposure should be made with 120 kilovoltage peak (kvp), 3.5 miliampere - seconds (mas), and a 72 inch source - to - image distance (sid). the room is small and the exposure must be made at 55 inches. how should the technique be changed?
increase mas to 4.1
increase mas to 4.7
decrease mas to 3.0
decrease mas to 2.0
Step1: Recall the inverse - square law for x - ray exposure
The relationship between the intensity of x - ray exposure and the source - to - image distance (SID) is given by the inverse - square law: $\frac{I_1}{I_2}=\frac{SID_2^{2}}{SID_1^{2}}$, where $I$ is the intensity of the x - ray exposure and SID is the source - to - image distance. Since the mAs is proportional to the intensity of the exposure to maintain the same image quality, we can also write $\frac{mAs_1}{mAs_2}=\frac{SID_1^{2}}{SID_2^{2}}$.
We are given that $mAs_1 = 3.5$, $SID_1=72$ inches, and $SID_2 = 55$ inches.
Step2: Rearrange the formula to solve for $mAs_2$
$mAs_2=mAs_1\times\frac{SID_2^{2}}{SID_1^{2}}$.
Substitute the given values: $mAs_2 = 3.5\times\frac{55^{2}}{72^{2}}=3.5\times\frac{3025}{5184}\approx3.5\times0.584=2.044\approx2.0$ (this is incorrect, the correct formula should be $mAs_2=mAs_1\times\frac{SID_1^{2}}{SID_2^{2}}$).
The correct calculation is $mAs_2 = 3.5\times\frac{72^{2}}{55^{2}}=3.5\times\frac{5184}{3025}\approx3.5\times1.714 = 4.799\approx4.7$.