QUESTION IMAGE
Question
iron(iii) thiocyanate spectroscopy
dilution process
the student creates 100.00 ml of each of
the dilute solutions from the concentrated
stock solution. below is a list of the steps
involved in the dilution process.
step 1: obtain a 100.00 ml volumetric
flask.
step 2: measure out 60 ml of 0.1000 m
fescn²⁺ using a beaker.
step 3: use a funnel to pour
the fescn²⁺ into the volumetric flask.
step 4: add distilled water to the fill line. cover
and allow the solution to mix.
which step in the dilution process is not
correct and why?
step 4: the distilled water should be measured
using a beaker.
step 2: you should use a graduated cylinder or
pipette for measuring the stock solution.
step 3: exactly 40.0 ml of water should be placed
into the volumetric flask first.
step 1: you should use a graduated cylinder for the
dilution.
- For preparing precise dilutions to a fixed volume (100.00 mL), accurate measurement of the stock solution is critical. A beaker is not designed for precise volume measurement; graduated cylinders or pipettes are the correct tools for this step.
- Analyzing other options: Step 1 is correct (volumetric flasks are for fixed-volume dilutions). Step 3 is incorrect logic (water does not need to be added first, and the volume is not fixed at 40 mL). Step 4 is incorrect logic (distilled water is added to the calibration mark, not measured in advance with a beaker).
Snap & solve any problem in the app
Get step-by-step solutions on Sovi AI
Photo-based solutions with guided steps
Explore more problems and detailed explanations
Step 2: You should use a graduated cylinder or pipette for measuring the stock solution.